  • 1.  Boston Scientific Lead Recall

    Posted 08-29-2025 14:23

    We are finding that we have several patients with a single spike in impedance, that comes back down to normal. What are you doing with those types of patients? With all of the ongoing advisories, how are you managing "advisory fatigue," so that in 6 months, a patient whose device has had normal impedances, who has increased impedance above the 150 ohm mark, does not get missed?



    ------------------------------
    Janet Powell MSN, FNP
    Nurse Practitioner
    Ballad Health CVA Heart Institute
    Kingsport TN
    (423) 230-5000
    ------------------------------


  • 2.  RE: Boston Scientific Lead Recall

    Posted 09-02-2025 09:44
    HI Janet, you raise an interesting point. The advisory recommends using
    the "most recent 28 day average impedance" to determine risk category.
    Since mathematically calculating the average is certainly not practical,
    one way to estimate this is by using the impedance trend graph.

    MURJ developed an advisory tool that displays all patients in the advisory
    with their most recent lead measurements as well as their most recent trend
    graphs. This visual is very helpful in managing these advisory patients,
    and the report can be run periodically to quickly assess for any patients
    who are moving into a higher risk category.

    There is also the Latitude alert for high impedance which should trigger a
    transmission and an audible alert. The Emory team is making it a habit to
    test the beeper in clinic for all BSX patients and making sure it is turned
    on.

    Hope this helps,

    Sallie

    --
    Sallie Gustafson RN, BSN, CCDS

    Director Medical Affairs

    MURJ, INC.